Hello Everyone! Long time reader of these posts and have gained much knowledge from everyone here.
We have a new to us wiped film machine in our lab and have been having issues with vacuum pressure. The unit we got is a HVE 30 with the booster pump kit as well as the beaker and wrench outlet pumps. Our feed tank is a peristaltic pump that feeds from a metal pot.
Terp strip has gone fine every-time we have done it and our parameters are below:
Feed: 100C
Body: 170C
Outlet pumps: 90C
Vacuum: stays around 0.3-0.5 torr
First pass seems to drop a lot more waste than we want and I think it is due to our vacuum not being able to get lower. Parameters below:
Feed: 100C
Body: 163C
Condenser: 90C
Outlet Pumps: 90C
Vacuum: can only get down to about 0.2 torr
I have tried to contact HVE and Lab Society but have had no luck with them. Our next step is to confirm that the vacuum pump itself is not messed up by segmenting the line and testing every part. Just want to see if there are any setting we could change that would fix this.
Blank off the vacuum and see how deep it can pull being directly hooked up to the vac meter. And what pump do they supply? There’s a chance that it’s not powerful enough, if it’s the stock/turnkey pump that comes with it.
Yes, if you get the good compression fitting type that I use on my skids (the same one cts uses on there pumps) you shouldn’t have a problem.
The spring that B and W uses are weak and it always has a problem sealing. Hate those dam pumps. I refuse to buy there stuff which is why I make my own