Thatâs actually a really solid setup youâve got â same solvent system and use of an internal standard like CBGV already removes a lot of the common sources of variability.
On your question about working in mg/L: that doesnât change the dilution logic at all â it actually fits perfectly with how Empower handles calibration.
Your standards define the curve directly in mg/L, and your samples are back-calculated into mg/L and then converted to % potency using your sample weight and total dilution factor. So the key is still making sure that your dilution factor reflects the entire path from original sample to final vial.
Where I think your main opportunity is now (especially since youâre already getting strong R² values) is tightening the relationship between:
- dilution strategy across matrices
- calibration weighting (likely 1/x² for your range)
- and internal standard recovery consistency
Using CBGV as an internal standard is a good approach â Iâd just confirm that its response is stable across your full concentration range and that itâs not introducing variability at the low end of your curve.
Since youâre based in Alabama, youâre likely aligning with typical US cannabis testing expectations, so getting within Âą2% of your third-party lab should be very achievable with a bit of calibration and processing refinement.
If youâd like, I can walk through one of your actual prep workflows and Empower processing methods with you and pinpoint exactly where the deviation is coming from â usually at this stage itâs something subtle in how dilution or weighting is being applied rather than the instrument itself.
Happy to take a closer look if that would help.
It is wonderful that we were able to bring these in line- and get your calibration and dilution handling consistent among all of your matrices as well.
Once you have the ability to run a few batches and be able to demonstrate to yourself that everything is stable-( especially compared to your 3rd party lab)- it would be invaluable to hear your thoughts on how the adjustments have performed for you and if you are now demonstrating this 2% agreement.
A brief comment on what it was like to work through the method refinement would also be greatly appreciated, if you feel comfortable- I will begin supporting several other labs experiencing the same type of HPLC/Empower calibration issues, and a hands-on account will help improve the process of supporting other customers.
Thanks again for your time and let me know if anything else arises- I will gladly follow up with further assistance.