Just put the globe valves on today and tested them. It’s so fucking sick!!!
FUCK SUBMERSIBLES!!!
Because I use like 4 lbs of salt with 3lbs of dry ice the submersibles just shit out after a while because of too much salt. This thing just says whatever and moves the water
If you can, only restrict the outlet and your pump will thank you. Restricting the inlet can cause cavitation from the low pressure in the impeller and it will potentially hurt the seals
It’s not really a big deal, certainly not a safety issue just a little pump general knowledge.
On another note, she RIPS. Only ran for about 15 minutes tonight, have a few good vacuum leaks to fix still but we clocked around 40 lph. We’ll see if it’ll do 60 when the leaks are fixed and the feed is preheated
The grey schedule 80 pipe overhead goes to our central chilled water plant which is 215 tons. It’s only a 1-1/2 pipe into the lab though so somewhere around 30-40 ton capacity split between this and the rest of the lab
Any pump is going to add heat to the working fluid, it’s an optimization. Typically centrifugal pumps are around 70% efficient (there’s a really wide range depending on a bunch of factors), which means 30% of the electrical draw of your pump might as well be a heater. Bigger pump means more heat but it also means better turbulent flow and delta-t which make for better heat transfer.
The good news is that we have 24 feet to the bottom of the next floor above that drop ceiling. The bad news is that the floor is 18" thick concrete…
On another note for those who may be watching, we kinda goofed. The vapor holdup is insane. I think I’m going to spin the whole unit so the vapor feeds from the top of the condenser which should have been a very obvious thing to do. Oh well.
Off topic one sec, “use a warehouse”… If agricultural engineers were as smart as that statement, Bill Gate would never worry about cows and meat again. No one would. Js.