7 minute protocol for testing 9 neutral cannabinoids on SRI 310-MM GC

So you’ve got In House analytics in the form of an SRI 310-MM GC and you want to test for a bunch of different things. But you’re impatient and hate waiting 14 minutes per test.

So how would you like to be able to quantify 9 cannabinoids in 7 minutes instead?

Here’s how we do it:


Sample Preparation
According to SRI standards, 100mg of sample in 40mL methanol with internal standard

GC Hardware Information
SRI 310C Gas Chromatograph
Restek MXT-502.2 5m, 0.53mm ID, 0.25um guard column followed by
Restex MXT-502.2 30m, 0.53mm ID, 0.25um column
On-column injection - [NOTE THAT COLUMN IS ON NON-STANDARD (terp/residual solvents) INJECTOR]

Restek Standard 34132
Cannabinoids Neutrals 9 Standard
1000ug/mL each in P&T Methanol:

Cannabichromene (CBC)
Cannabicyclol (CBL)
Cannabidiol (CBD)
Cannabidivarin (CBDV)
Cannabigerol (CBG)
Cannabinol (CBN)
d8-Tetrahydrocannabinol (d8-THC)
d9-Tetrahydrocannabinol (d9-THC)
Tetrahydrocannabivarin (THCV)

GC Events Program
0.000 Zero
0.000 Sound
2.000 Integration - Based
2.500 Integration - Based
3.000 Integration - Based
3.500 Integration - Based
4.000 Integration - Based
4.500 Integration - Based
5.000 Integration - Based
5.500 Integration - Based
6.000 Integration - Based
6.980 Sound

GC Temperature Program
Initial Temp | Hold | Ramp | Final Temp
200 | 0 | 25 | 250
250 | 0 | 15 | 310
140 | 1 | 0 | 140

Integration
Peak: 90.00%
Base line: 75.00%
Area reject: 10.000
Standard weight: 4.000
Sample weight: 100.000

Here’s what a chromatogram from one of our calibration runs looks like:

We only calibrate for CBD, THC, CBG, and CBN as that’s all we presently care about.

Happy testing.

19 Likes

and that is CBD and CBC that you’re not getting baseline separation on?

elution: THCV, CBDV, CBL, CBD, CBC, D8,D9, GBG, CBN ?!?

7 Likes

That sounds correct. I’ll try to remember to check the peak order next time I’m in the analytical lab, I believe we’ve got a chromatogram up on the wall with each peak labelled.

Now that we’ve got the HPLC up and running the GC mostly sees work doing residual solvent testing.

5 Likes

How is the 310 going for you a year later?

1 Like

We sold it as it got very little use after we got our HPLC dialled in.

4 Likes