Who sells huge cryo chillers that aren't huge disappointments?

@precisionnick

1 Like

We have bigger chillers which run 24/7 in Pharma , Aerospace and Automotive.
Most of them critical applications. We are market leader with these systems.
Greenbuggy is undersized big time and of course our 815 has a struggle with 1.5 KW where at least 4 KW is needed.

5 Likes

Well that got interesting.

4 Likes

@huber_king_of_temp: thanks for stoppping by and chiming in.

Any chance you could lay down the sizing math for us?

Or just confirm that this is close?

    Ethanol = 46g/mol

    has a heat capacity of 112J/mol.K

    and a density of 0.7893g/cm3

    1gal = 3785.41cm3 (mls)

    => density = 2990g/gal

    => 100gal = 299000g

    => 299000g/46g/mol = 6500mol.

    to lower  6500mol by 40 degrees C (or Kelvin) would be 
    40 * 6500 * 112 = 29,120,000J = 29,120kJ

The above tells me how much heat I need to remove. It doesn’t tell me how big a chiller I need to accomplish this. What else does one need to actually size the chiller correctly?

3 Likes

Hello Cyclopath

I am happy to give you some guidance >
first of all I need all masses ,
what is your reactor material : glass or stainless ?
What is the approx. weight ?
What is the reactor volume ?
Reactor content is ethanol ,
What is your starting temp ?
What is your goal temp ?
What is your time to goal temp expectation ?
If I have all this information I can calculated the chiller kw you need

4 Likes

Doh! don’t want to ignore those. Vacuum jacketed stainless.
Don’t have a clue atm. but understand that like the solvent they have a heat capacity and will need to come along for the temp drop. call it 200lb at a guess.

Math shown was for 100gal.
It was done for 20C => -20C
but that doesn’t change the energy that needs to be removed, just how hard it is to remove it :wink:

lets start at ambient. and lets assume that we’re in a poorly controlled warehouse in summer…so that actually means 35C.

Because it happens.
It’s also my expected solvent boiling temp in the recovery still :wink:

lets go to -60C. because it’s hard. but looks like it might be the right response.

how fast? I gave a rational for wanting to get from 20 to -20 in a hr over here…suggesting I need to go from +35 to -60 in that time-frame without throwing liqN2 at the problem seems like asking for a BIG chiller…but one hr temp on a 100gal of solvent would match up nicely with 100gal/hr folks would like to see out of their solvent recovery rigs. So lets call it one hr (also allows scaling to 1600gal “over-night”).

Can I convince you to show your work?

If you teach the village how to size their chillers, or at least double check the consultant they hired, there won’t be as many folks like the OP running woefully undersized chillers…

7 Likes

yep , that is exactly right …jacketed reactor material , hoses, and reactor content all have different heat capacities. that’s why we all take them into consideration when calculating heat removal
Then we take a starting temp 20 C and a goal temp -60 C, that’s a delta of 80 K and lets assume we want to reach our goal temp in 1 h…that would require approx. 15 KW of cooling power in 35C ambient I would add another 20% so with 18 KW it should be easily accomplished. For this big sizes air cooled is out of the questions since the heat output is way to high , so that leads to water cooled chillers.
If you allow 2 h of cooling down time chiller size is about half…
We have a great source of case studies and here is one on a 250 Liter ( 65 Gallon ) Stainless Steel reactor.
http://www.huber-usa.com/download/casestudies/CS_74.pdf
If you need more help sizing chillers Please contact info@huber-usa.com anytime , we are happy to help
Hope this guidline helps the community to make the right choices

12 Likes

my response above :slight_smile:

2 Likes

Haha, yes, I was going to say, no way a single chiller of that size can cool 105 gallons of alcohol in an hour to that target temp.

People who don’t have knowledge of thermodynamics should not build physics labs. There are plenty of knowledgeable consultants out here that can help avoid these mistakes.

5 Likes

what is really needed is some way to change the physics of ethanol recovery.

which is what the CXE rabbit hole is all about I believe :wink:

3 Likes

you are talking my language SoStupendous

1 Like

way undersized is not a question of quality

1 Like

I’m well aware that the Unistat 815’s are nowhere near large enough in capacity to do 105 gallons of ethanol in an hour. The 105 gal/hr figure is what I’d like to be doing short term, and would like to double that in the longer term. I know that means spending more money and buying more equipment. The lack of capacity is only a portion of my complaints about these units.

I’ll note that I was NOT the person who spec’ed out these chillers. I don’t know if it was that person who didn’t (or did wrong) the math to calculate what kW chilling capacity they were going to need, or if they got taken by a salespersons claim without the technical understanding to disprove the sales-speak. I’m just trying to do the best I can as far as production with what I have available to me at the current time. I know full well that some capital expenditure is inevitable if I want to increase my production rate, right now the current rate is not even close to what we need to achieve to make money.

I received a call from Huber this morning and it sounds like they are going to work to get the technical issues with these machines taken care of (pump issue and touchscreen issue). They apologized that nobody had gotten back to me about my inquiry with regards to their Spycontrol software that isn’t working and said they would work on that as well, to the extent that they’re willing to fly out a tech to address all these issues if need be.

They said that for the size of reactors we are running these Unistats are undersized and for the volume of liquid being circulated should have been fitted with an auxiliary expansion tank. I don’t fault Huber for pointing these things out - though I think there’s a good possibility that the dealer who sold our company these knowing full well what our needs are may need additional training to help customers get the right equipment the first time.

I will keep this thread updated as Huber works with us to resolve these problems.

5 Likes

I run three water cooled three phase julabo FPW-91SL’s and have had issue after issue. Have heard good and bad about huber but at the scale you’re talking I would look towards industrial applications…

1 Like

Hi Derek…
using open bath systems down to -90 is always a challenge…
if you like to contact me I would love to help you .
We have 6 service technicians across the country , so our support is much better then most other competitors

2 Likes

We are using an iced tech chiller right now and I will vouch for it working pretty well. Keeps up with a giant crude extractor using two master vapor pumps right now. Still not quite getting the recovery rates they would tout in the demos but it’s been pretty reliable the last month or two.

The cons: Yes, you have to put it on the roof, and it’s probably not moving once it’s installed. Very difficult to adjust the temperature setting once it’s installed. Basically has one temperature - fucking cold.

4 Likes

Ive been to your operation, your collection pots are those massive cube tanks that feed The King of Rock. 3 huber 815s are simply to small. They are great for holding decently sized vessels at low temps for long time, or keeping up with medium depth condensing. Trying to chill that much solvent that quickly takes a ton of power. I would augement your cooling with liquid co2 in coils or in jacket if said reactors if applicable. Youll need a large co2 tank outside to have enough stock up last a while.
But just dropping dry ice in your ethanol is a joke for anything over benchtop scale. Not only does it cause misting and external ethanol condensation, it will leach acids into your solution and could cause further isomerization down the road which will kill isolate yields and potency reaults.

6 Likes

Yep, and Monday we’re going to sit down with Huber and figure out how to move forward with our operation. We aren’t keeping the massive (375 gal) cube tanks cold with the Hubers (once plant material is separated from alcohol no point in keeping it cold), we’re just trying to keep the 35 gal jacketed reactors cold and they aren’t keeping up with that either.

2 Likes

The King of Rock? :swoon: :heart_eyes:

3 Likes

I have calculated the cooling power for about 100 Gallons of Ethanol from -25 to -45 C
Huge is a relative term.
We have a industrial size Unistat 630W ( Big Big brother of Unistat 815 )
14 KW @ -40 C / 5 KW @ - 60 C…Listprice 176,510
This machine can do this job easy . CO2 is much more complex , and expensive.